10-29-2013 02:16 AM
Hi, we are able to communicate the Sorensen XG 300-2.8 from Hyperterminal but when I try to communicate with the MAX -> Serial Interface -> Open visa panel. I can write to it but I couldn`t neither read or query as we are getting error. We are communicating through USB, Unfortunately we couldn`t communicate through MAX as well as through LabVIEW. We have installed all the drivers required please suggest some solutions or guide us in troubleshooting the issue or may give advise to what to look for?
10-29-2013 02:23 AM
Hello samputhur,
What errors are you getting? Have you attempted to use the LabVIEW instrument drivers provided here:
NI Instrument Driver Network: Sorensen XG Series
http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=20077
Error codes and screenshots of the errors you're seeing are extremely helpful.
Regards,
10-29-2013 04:03 AM
Dear Mr. Tom,
Attached are the error code and the error message we encountered. We had already installed the IVI and LabVIEW drivers as you have mentioned We got the error while trying to initialize the sorensen power supply using PLP drivers as seen in the image and the other error message while trying to read the status bit from MAX. Please let us know what you think. We look forward to hearing from you.
Best Regards,
Sam Puthur
10-29-2013 12:06 PM
10-30-2013 01:39 AM
Hello Dennis,
In one of the earlier forums you have mentioned about the drivers you created by yourself for XG 600 series can you please share the driver files http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/Can-t-communicate-with-Sorensen-XG-series-thr...
from the comments by CHKNRCKD from the link we tried in MAX, the commands default termination character to be xD, and added termination character buffer to be *IDN?\r\n.
But while reading i`m still getting the error 0xBFFF0015.
Attached is the attribute error from MAX since we are fairly new we don`t know what to interpret out of this.
10-30-2013 11:48 AM