I have tried to get 6 thermocouple reading using Niswitch Thermocouple measurement.vi (hardwares are PXI2527 &PXI4071). When I run the program with scan list Ch0:5->com0; the program is working but the relays are frequently switching On &Off. I am wondering it would reduce the lifetime of relays.How can I get the data with holding up the relays? I am currently using labview 8.5.Thanks
06-27-2013 06:48 AM
According to the specifications the mechanical lifetime of the relay is 1E8 cycles but the electrical life at low current is 2.5E6 cycles. That is about one month at 1 cycle per second.
There is also a calculation of the errors you can expect when using thermocouples.
Lynn
06-27-2013 08:18 AM
06-27-2013 07:44 PM
Hi Riyad12,
The DMM only has one channel, so only one measurement can be taken at a time.
You can set the Scan Delay property to add a pause between execution of scan list entries. As you put it, this will hold the relays in the connected state for a specified period of time.