Type K (Chromel / Alumel)So, at 100°C, without any amplification, the thermocouple would output a voltage of
Type K is the 'general purpose' thermocouple. It is low cost and, owing to its popularity, it is available in a wide variety of probes. Thermocouples are available in the -200 °C to +1200 °C range. Sensitivity is approx 41 µV/°C. Use type K unless you have a good reason not to.
.000041V/°C * 300°C = 0.0123V (the TC range starts at -200, so add 200 to the desired temp)And at 25°C, the TC output would be
.000041V/°C * 225°C = 0.009225V
10-30-2007 11:40 PM
10-31-2007
11:34 AM
- last edited on
04-10-2024
02:15 PM
by
Content Cleaner
Hi Alexx631,
The scaling will take into account the thermocouple gain and the copper-thermocouple junction temperature.
If you use an example such as Acq Thermocouple Sample from the NI Example Finder then you can see they
create a Analog Input Temperature Thermocouple task. This tells DAQmx to perform the scaling and take into
account the CJC temperature.
The CJC can be a constant or taken from a temperature sensor which is located physically at the Thermocouple/Breakout box junction.
The SCB-68 and SCC-68 are example's of NI breakout box which incorporates a temperature sensor for built in CJC.
In your case, it sounds your performing a analog input voltage reading. This will return the raw voltage output of
the thermocouple + the voltage offset generated by the thermocouple and copper junction.
I would suggest taking a look at the Acq Thermocouple Sample example from the NI Example Finder.
Thanks, please do post back if you still have questions.
Best regards,
MatthewW
Applications Engineer