Timeout due to less amount of buffer size

Solved!
Go to solution

A few things.

 

1. What error are you expecting?   Much of the discussion has been about the way DAQmx Read *allows* you to request more samples than the buffer size such that you don't *need* to expect an error.  Please reread responses in msg #2 and msg #5 that provide more detail.

 

2. Simulated devices are not (and were never intended to be) *perfect* simulations of the behavior of a real device.  They do a rather good job of catching problems with configuration syntax, settings limits, and device capabilities.  Barring that, simulated reads and writes pretty much always succeed.

    With a real device, I very much expect you'd need a buffer size larger than 10 when you sample at 250k.   A 250k task would fill a size 10 buffer every 40 microseconds.  It's extraordinarily unlikely that application-level code would be able to service that buffer frequently enough to prevent buffer overflow.

    So in this kind of extreme case with a really tiny buffer, I guess you might expect an error on real hardware that the simulated device didn't anticipate.  But it's a pretty artificial situation.  You have to go out of your way to make the buffer size so small, much smaller than you'd get with DAQmx's buffer auto-sizing.

 

3. You referred to a "daq write" this time.  Aren't we talking about an acquisition task and DAQmx Read?   Behavior for writes is different as I mentioned in msg #9, and different in a way that might show up with a simulated device.

 

 

-Kevin P

 

ALERT! LabVIEW's subscription-only policy came to an end (finally!). Unfortunately, pricing favors the captured and committed over new adopters -- so tread carefully.
0 Kudos