Hello. We are hoping to control a detector via NI PCI-6533 (DIO-32HS). Our colleagues have implemented the control programs in LV 7.1 with Traditional NI-DAQ 7.3. Now when we opened their programs on a PC installed with LV 8.6 and NI-DAQmx 8.9, many vi's were not located, including "DIO config.vi", "DIO wait.vi", "DIO clear.vi", "DIO read.vi", "Device reset.vi", etc. We realized that they were supposed to be from vi.lib\daq\digital.llb and some other .llb in the same folder. Apparently vi.llb\daq was installed by the traditional DAQ.
We don't think there is traditional DAQ compatible with LV 8.6, which means we may need to replace those "DIO *.vi" or "Device reset.vi" with their counterparts in DAQmx. Could anyone throw some light on how we should start working on this ?
Thank you very much.
Solved! Go to Solution.
04-09-2009 03:12 PM
04-10-2009 10:46 AM
Xiaolan,
There is indeed a Traditional DAQ driver for LV 8.6. The install can be found here. Hope this helps.