Dear Labview community,
I need help as regards to my assembly for data collection.
I have for materials :
- a triaxial DC response accelerometer model 3713B1110G;
- a 3 channel DC sensor signal conditionner model 478B05;
I have already connected the two devices thanks to a wire with 3 inputs.
Now, I need to link my assembly (accelerometer and conditionner) to my national instruments DAQ device (NI USB6225).
Do you know what type of wires I need?
Please find in attachments the model of DAQ and the conditionner I would like to link.
Best regards
Solved! Go to Solution.
05-30-2017 12:28 PM - edited 05-30-2017 12:33 PM
the output of the conditioner is BNC. I assume, but I'm not shure that all BNC shields share one common potential (check: measure the resistance between them with a DMM or RTFM)
Since you have enough differential channels: Use them and configure your input task to be differential.
connect each conditioner output x to AI{x} (inner pin) and AI{x+8} (shield) (x from 0..2 😉 )
If all conditioner shields are same common potential, add a 10k to 1M resistor from one of them to AI GND, or 3 for each otherwise. (you can do that in the connection box there are a lot of AI GND connections 😄 )
An easy way to get connection cables: BNC to BNC cables a usually easy to get. Grab two of them with double length needed and cut them in the middle 😛
professional next steps involve wire end ferrules and shrink tubing 😉
and think of the strain/pull relief
05-30-2017 12:45 PM
thanks a lot for your precious help! It is so nice!
Greetings from New York
Fanny S