11-12-2009 11:21 PM
All most certainly due to EMI coming from the drive. The drive almost certainly has a base frequency of 3 -6 kHz. And it can radiate that strongly enough to interfere with other cabling.
The first thing I would check is to put an oscilloscope on the analog wires going to your USB-6218 and see if you get any high frequency noise on that.
I'm not sure why LabVIEW would freeze, unless the DAQ functions you use are waiting for data from the USB device that has disappeared and all other code in you program is waiting for the data to be returned from the DAQmx functions. That or the noise coming in from the USB cable is not just screwing up the reading from the DAQ device, but also the basic operation of the PC. Does everything else on the PC seem to run normally while LabVIEW appears frozen?
11-13-2009 08:15 AM
I was incorrect in saying Labview freezes. The VI that is reading from the USB-6218 goes haywire. I recieve all NAN on my anolog inputs and the code becomes extremely sluggish. The easiest thing to do is kill the VI with ALT+CRLT+DEL.
I failed to mention another detail. I am trying to run a second VI at the same time. This VI controls the Seib & Meyer motor drive via serial port communication. This VI does not have any problems when the USB-6218 goes haywire. We have tried running the serial communication code from a second PC to determine which VI has the problem. The problem only occurs when both the USB-6218 and the Serial code are running from the same PC. The Serial communication does not exhibit any problems when the USB-6218 goes haywire.
I have run serial communication VI's and DAQ VI's simulaneuously before. I only have this problem with i combine all 3; serial communication, USB-6218 and the high freq motor drive.
11-13-2009 09:02 AM
Like Ravens Fan, I suspect that your problem is EMI or possibly a ground loop. The USB-6218 is isolated. I think that means that the signal inputs are isolated from the USB ground. If your 32 kW motor drive is inducing significant currents (perhaps on the order to 10s to 100s of mA) into the USB cable, the OS could lose communications with the device.
You indicate that you are getting NaN from the device. That suggests that the subVI is still running. Are you getting any error messages or timeouts? Specifically what error messages? Have you put any indicators on loop counters or state selector lines to show what is happening in your code when it gets sluggish? What is your code architecture?
Lynn
11-13-2009 10:32 AM
11-13-2009 11:54 AM
Actually, your code looks quite good. There are a few things that could be improved.
1. Replace the old file open/close functions with the new versions. If you do that and a little bit of rewiring, you can completely eliminate the stacked sequence structure.
2. The index array function is expandable, so you don't need two or more of them to get out multiple elements.
3. In the DAQ VI, you could wrap up the waveform manipulations and Mean calculations in a For Loop and use auto indexing on the output. If you reorder you DAQ task just right, you could easily do that and order the outputs in the same order that your later graph and write to file can use.
Back to the problem at hand:
You said, "The problem only occurs when both the USB-6218 and the Serial code are running from the same PC. The Serial communication does not exhibit any problems when the USB-6218 goes haywire." So if the serial code is running on a separate PC from the DAQ code, then the DAQ code runs okay?
I think you'll definitely need to do what you can to move and isolate the USB DAQ device. Perhaps even mount it in its own metal box and ground the metal box. Is the motor controller in the open or in its own control panel box?
If you find that the serial communication is somehow affecting or being affected by the EMI, then try putting an isolator on the serial wiring. I would also recommend lowering the timeout on your serial communication from 10 seconds to something like 2-3 seconds. Basically just a little bit longer than whatever your expected time is between when you initiate a read and expect to get the data back. No reason to wait any longer than necessary for data that is not coming. On your DAQ side, you may want to lower the timeout value as well from the default 10 seconds. It won't work any better, but it will come back quicker with an error if the USB gets lost and not seem as sluggish. Also put an indicator on your Error Out wire from the DAQ assistant.