Dear Support,
I'm using an OEM 6501 USB Device from NI, at power up prior to any LabVIEW commands, digital output level is HIGH ( open drain with internal 4K7 pull-up resistor). Unfortunatly, my application needs a LOW level at default, so i have to replace Open Drain configuration by push-pull and i would like that OEM USB 6501 device starts automatically with a low level.
How to proceed to configure Push-pull mode to replace Open Drain default mode with LabVIEW commands?
Do i need to suppress all 4K7 pull up resistors on the board?
Thanks for your help.
Regards,
Marc,
11-24-2025 06:29 AM - edited 11-24-2025 06:30 AM
Hi marcounet,
@marcounet wrote:
How to proceed to configure Push-pull mode to replace Open Drain default mode with LabVIEW commands?
There's no way to change the power-on state, your device acts exactly as described in the specs: "input (high-impedance)" on startup!
@marcounet wrote:
Unfortunatly, my application needs a LOW level at default,
Options:
@marcounet wrote:
Dear Support,
This is a public forum, mainly driven by volunteers.
When you want to contact the NI support you need to start a service request within your user account…
11-24-2025 06:34 AM - edited 11-24-2025 06:41 AM
The 6501 I/O pins do not provide power-up default settings.
See Re: Can I change the power on state of the USB-6501? - NI Community
There are other models (USB-6509) that support power-up defaults.
11-24-2025 06:43 AM
Thanks i don't know if USB6509 exist in OEM version.
Regards,
Marc,