Unwanted gain created by AFG3022 Function Generator Instrument driver Example VI

SOLVED
SRoy6
Member
Solved!

Unwanted gain created by AFG3022 Function Generator Instrument driver Example VI

I was using the 'Create Standard waveform' example VI of AFG3022 function generator. I was able to create a waveform but the gain of the waveform looks 5 times less on a scope. For example, if I create a 5 Vpp square wave of 150kHz through the example VI, it looks like a 1Vpp, 150kHz square wave on my scope. Why does this happen and how can I get rid of the gain?

3 REPLIES 3
kbbersch
Active Participant
Solution

Re: Unwanted gain created by AFG3022 Function Generator Instrument driver Example VI

Does the impedance setting for your scope input match the probe impedance?  I've seen this when the input impedance was wrong on the scope.

 

Kelly Bersch
Certified LabVIEW Developer
Kudos are always welcome
Dennis_Knutson
Knight of NI

Re: Unwanted gain created by AFG3022 Function Generator Instrument driver Example VI

And what is the output of the 3022? I suspect it's 50 ohms so obviously the scope will read a different amplitude. It doesn't sound like you have a LabVIEW problem at all. You can set the amplitude manually to verify it has nothing to do with any code.
SRoy6
Member

Re: Unwanted gain created by AFG3022 Function Generator Instrument driver Example VI

You're right ... It was the impedance setting , but on the probe which was messing up my measurements!