I just installed and configured GPIB-ENET/100. The troubleshooting wizard says it's functioning properly. After running a Labview Program, I get VISA error: 1073807304 occurred at VISA write in AW372XXA INIT open vna.vi -> xxx.vi VISA: (HEX 0XBFFF0038) bus error occurred during transfer. What can I do to fix this?
That corresponds to the NI-488 error EBUS, which is usually only returned if you do not have a cable connected to your GPIB controller (ie, the GPIB-ENET/100). If you do have a cable connected, then the cable may be faulty - try another if you have one.
Dan Mondrik Senior Software Engineer, NI-VISA National Instruments
I want to measure the C-V using Agilent e4980a. Please I need a program labview 8.5 for the manipulation, and it can help me to find the labview 8.5 program adequate. vi thank you