Voltage measurement not proper on NI9201
I made a simple program to measure DC Voltage using DAQ Assist. The problem i am facing is that when i connect no power supply to DAQ it is not showing 0 voltage and it is fluctuating due to which I am not able to take proper readings
(1,192 Views)
Hi Rohit,
The NI9201 module doesn't require any external power supply (supply on its own): what are you talking about?
(Ofcourse you need to power the chassis where the NI9201 is plugged in!)
When the chassis is not powered then you should get a DAQmx error…
You can make "proper readings" even when the readings are fluctuating!
(1,188 Views)
Hii thanks for replying
i was talking about the voltage values measured by the module. I wanted to measure voltage of a device but when I ran the program without connected any device. The program was not showing 0 voltage and it varied highly
(1,183 Views)
What are you reading in NI max with nothing connected?
Sounds like dodgy grounding or ghosting.
(1,171 Views)
What is the input configuration?  I suspect that you have set to NRSE (non-referenced single ended).  With neither signal source nor reference connected,  what you describe would be expected.  What happens when you read a RSE measurement or connect the reference to ground?
(1,150 Views)
NI does not specify what voltage a device will read when there is no signal connected.
Reference: Incorrect Readings on Unconnected or Open Channels of DAQ Device
https://tmesystems.net/