Voltage spikes using USB-6009

Many multichannel data acquisition devices multiplex several inputs into one A/D converter.  Both the converter and the multiplexer may be switched capacitor devices.  Such devices require input current to charge the capacitors and experience large, fast changes in the instantaneous input impedance when switching.  You may see an improvement by putting a linear differential amplifier, such as an instrumentation amplifier, between your transducer and the DAQ device.

 

What is the output impedance of the anemometer?

 

Lynn

0 Kudos
Message 11 of 21
(1,386 Views)

Thanks for that.  The output impedance of the CTA is 50Ohms as specified in the manual.

 

I have actually tried experimenting with the gain settings built into the CTA device which possibly reduced the spikes for zero flow (or rather there influence) but once there was flow on the wire the anomalous spikes were as before.

---------------------------------------------------

LabVIEW 8.5 User.
0 Kudos
Message 12 of 21
(1,384 Views)

With a 50 ohm output impedance you should be fine.

 

I suspect that the problem is something else but that the symptom shows up as a result of the overall configuration or something subtle.  That is why I asked about what you saw with the scope.

 

How is everything grounded?  Do you have resistors to ground on both of the differential inputs to the 6009?  How is the motor for the wind tunnel controlled?

 

Lynn

0 Kudos
Message 13 of 21
(1,372 Views)

The motor for the wind tunnel has its own separate supply.  The blower is around 3m in diameter and has a lot of inertia so I can actually cut the supply and still have significant flow, upon which I'm still seeing the spikes.  I know the tunnel has previously been used successfully (albeit with different hardware).

 

I only have one resistor from -ve to ground.  I have read through this link ( http://zone.ni.com/devzone/cda/tut/p/id/3344 ) regarding field wiring and noise considerations for analogue signals and believe this to be adequate as I want to run DC-coupled, please correct me if I'm wrong here.

 

Thanks.

---------------------------------------------------

LabVIEW 8.5 User.
Message 14 of 21
(1,364 Views)

Look at Figure 11 and related text from that document.

 

Lynn

0 Kudos
Message 15 of 21
(1,362 Views)

You reckon I should have two resistors tied to ground, where R1=R2?  Why would that be, because the source impedance is relatively high?

 

I can give it a try.

---------------------------------------------------

LabVIEW 8.5 User.
0 Kudos
Message 16 of 21
(1,357 Views)

No.  It has nothing to do with source resistance.  It is a path for the amplifier's bias current.

 

Lynn

0 Kudos
Message 17 of 21
(1,352 Views)

Could this explain what it is that I'm seeing with the voltage spikes?

 

Wouldn't that result in the mean voltage being brought down to 0V, as in AC coupling?

 

Apologies for all the questions!  I'll try this tomorrow with two 10k resistors.

---------------------------------------------------

LabVIEW 8.5 User.
0 Kudos
Message 18 of 21
(1,349 Views)

It is not clear that this is related to the spikes.  However the input stages of the multiplexer and converter may do strange things when not terminated properly.

 

With the 50 ohm output impedance of the anemometer 10000 ohm resistors are probably OK.  Any mismatch in the resistances can introduce errors and both the common mode and differential input impedances of the system are affected by these resistors.  Unless the specifications for your device say otherwise I would probably start with 1000000 ohm resistors.

 

Lynn

0 Kudos
Message 19 of 21
(1,338 Views)

sample.PNG

This is the input circuit of the daq 6009 board. I know the daq 6009 is low cost. But still the input stage is very clumsy designed. As you can see resistor are already in place.

What is the noise in your system with only a battery connected to daq unit. It could be nice to know that.



Besides which, my opinion is that Express VIs Carthage must be destroyed deleted