06-23-2010 02:19 PM - edited 06-23-2010 02:20 PM
Hey Everyone,
I am currently working with 3 different Watlow F4 Controllers attached on environmental chambers. I have developed a full user interface in order to create a profile, save it and load it into the oven, and gather information off the oven such as current temperatures, humidity levels, step number, etc... We have 2 Single chamber ovens and a dual chamber oven. My issue that I am encountering is that on the single channel ovens, when I set the Ramp Rate, the oven will randomly soak down to a temperature of -38C for a minute then execute the Ramp Rate at a much longer set time. On the dual channels, the Ramp Time is suppose to take 1 hour to ramp from 80C to 30C but it is taking 9 hours. Has anybody run into such an issue or know of a fix for it? The ramp rate for the single channel will work most of the time but randomly will randomly throw a soak into the profile and the ramp time for the dual channels will work on the rise but not on the temperature drop. Any ideas??
06-23-2010 02:30 PM
@Fonzie1104 wrote:
Hey Everyone,
I am currently working with 3 different Watlow F4 Controllers attached on environmental chambers. ...
. Any ideas??
Only one based on experience with a Cincinati Sub-Zero chamber with the same controller.
Make sure the PID is tuned correctly!
I spent four days on-site years ago watching out of control temperatures before I got the PID tuned correctly for the chamber.
Ben
06-29-2010 09:46 AM
What is the firmware version on this F4? This can be viewed when the F4 is first powered up, and is displayed for a few seconds on the lower display. If it is version 20.0, then I recommend that this firmware be upgraded on the F4. To accomplish this you have a two options.
1. You can upgrade it by sending it back to Watlow via an RMA. If it is less than 3 years, then this will be covered under warranty.
2. You can upgrade to Rev 22.0, by using a zip, that we can send to you. Please read the file and follow the instructions per, F4 Flash Update Procedure.pdf; which is also included in this zip file.
If you have further questions please call 507-494-5656, Watlow Tech Support line; and/or email wintechsupport.com
06-29-2010 10:06 AM
Could you post your code?
We have a thermal chamber with that same controller here and I could run it and see what is goign on.
06-29-2010 10:09 AM
I tried to post it, but the size was 9 Meg, whereas the acceptable maximum size is 5 Meg.
If you can give me your email address, I can email you the file.
06-29-2010 10:14 AM
The two that I tried my program on was a single channel (Rev. 2.06) and a dual channel (Rev. 2.09). They both have the ramping issue that I posted above where the single channel ramps to -38C randomly and adds random soaks, and the dual channel does not ramp at the correct amount of time but soaks for the right time.
06-29-2010 10:17 AM
Email is posted in my profile
06-29-2010 10:43 AM
I could not see your email in your profile. Therefore I am sending you the link to the F4 Flash Update. Here it is: http://www.watlow.com/downloads/en/software/f4%20flash%20update.zip .
Since you have 2.06 & 2.09; the latest upgrade for your F4 is Rev 11.00. Again please read the F4 Flash Procedure pdf.
06-29-2010 11:04 AM
Thanks! Will this work with a GPIB connection?
06-29-2010 11:06 AM
No, it needs to update via the RS-232 connection on the F4.