Why am I getting intermittant time outs with my test set?

Here is the whole deal.

I have LabVIEW set up to drive an O-Scope and a DMM. The DMM is set up so that a GPIB Read will get a sample of the voltage. The O-Scope is set to trigger and capture 4 channels of data. So much for hardware.

The software is written in two "threads" in the same VI. The top one polls the O-Scope to see if it is running (waiting for a trigger) or has stopped (a trigger occurred.) If the scope has triggered, then the wave forms are sucked off the scope and written to disk. The bottom thread reads the next voltage from the DMM. When it gets 180 samples it writes the min and max to a log file.

Intermittantly the O-Scope thread gets a timeout error. A retry corrects the issue and by loo
king at error logs and log file dates it seems that the failure always occurs right before a new set of waveforms is written to disk. (Within a second.)

Two possibilities occur to me. The Code polls the O-Scope Right as it is capturing data and the query is lost or ignored. The other is that since both instruments are accessed asynchronously that somehow the query of the DMM sometimes screws up the query of the O-Scope.

It could be something else.

Any advice is appreciated.

Thanks!
0 Kudos
Message 1 of 7
(3,278 Views)
Hello,
The best thing to do is get an NI-SPY capture to see exactly what the sequence of the commands are. It is possible that the o-scope ignores the commands under certain circumstances. That is not expected behvaior on an instrument, but you should definitely contact your instrument manufacturer to see if that is a possibility. The DMM and O-Scope queries would not interfere with each other, your GPIB controller regulates who can send messages and when they can be sent. Only 1 message gets sent on the bus at a time (only one "talker" on the bus at a time).

You wrote "A retry corrects the issue" does the "retry" entail another just another read, or another write and a read.

Ray K
NI Applications Engineer

NI Spy Tutorial
0 Kudos
Message 2 of 7
(3,278 Views)
I would say you first assumption is probably correct, that the scope
has trouble handling the status query and data aquistion at the same
time. Is your timeout error generated by a GPIB read function or a
GPIB write function? If it is a read error, most likely the scope
must first handle the acquired data before it can respond to any
commands in the buffer. Have you tried simply increasing the timeout
value of the function generating the error (or if you're using VISA,
set the timeout with a property node)?

If a write command is generating the error, the scope is probably
asserting the NRFD (not ready for data) line on the GPIB bus. In this
instance, the write function will be instructed by the driver to halt
sending characters until the line is clear. You need to increase your
timeout value to give the write function more time to wait until the
line is clear.

I find the best way to avoid these situations is to take more direct
control of the instrument in question. It sounds as if the scope is
configured to "trigger" or "sweep" continuously, and you are just
saving the data if there is anything new. It may be better to trigger
the instrument to sweep from your program rather than letting it run
continuously. That way you have control over the sequence of events.

I use a few different brands of optical spectrum analyzers and they
all work better this way. First I issue the command to sweep (acquire
new data) then (depending on the model) I either wait in a loop while
polling the status byte to see if the sweep is done, or I issue the
*OPC? command, which causes the instrument to return a "1" byte when
the acquistion is complete. In the second method, I set the labview
timeout of the GPIB Read (the instance that reads the response of the
*OPC? command) to a value longer than the longest known sweep time for
the instrument. This insures the program will wait until the
instrument responds to the *OPC? command or until the timeout expires
(something went wrong). When using the first method of status
polling, it is sometimes necessary to wait a short period (a few
hundred milliseconds or so) to allow the instrument to update the
status byte with the "busy" status after acquisition has been started.
I found this to be true with Ando OSA's.

Finally, some instruments refuse to continue GPIB operation if the
commands you are sending are generating an error and the error message
buffer is full. This is pretty rare, but I have seen it. Check the
command for reading the error buffer and add it to your program
(:SYST:ERR? is common) to see if you are generating errors on each
iteration of the loop. Hope this helps.

gm




Prog Guy wrote in message news:<50650000000800000098610000-1031838699000@exchange.ni.com>...
> Here is the whole deal.
>
> I have LabVIEW set up to drive an O-Scope and a DMM. The DMM is set up
> so that a GPIB Read will get a sample of the voltage. The O-Scope is
> set to trigger and capture 4 channels of data. So much for hardware.
>
> The software is written in two "threads" in the same VI. The top one
> polls the O-Scope to see if it is running (waiting for a trigger) or
> has stopped (a trigger occurred.) If the scope has triggered, then the
> wave forms are sucked off the scope and written to disk. The bottom
> thread reads the next voltage from the DMM. When it gets 180 samples
> it writes the min and max to a log file.
>
> Intermittantly the O-Scope thread gets a timeout error. A retry
> corrects the issue and by looking at error logs and log file dates it
> seems that the failure always occurs right before a new set of
> waveforms is written to disk. (Within a second.)
>
> Two possibilities occur to me. The Code polls the O-Scope Right as it
> is capturing data and the query is lost or ignored. The other is that
> since both instruments are accessed asynchronously that somehow the
> query of the DMM sometimes screws up the query of the O-Scope.
>
> It could be something else.
>
> Any advice is appreciated.
>
> Thanks!
0 Kudos
Message 3 of 7
(3,278 Views)
The retry retries the entire sub VI, and three different sub VIs have the retry ability. The Sub VI that I see failing write a querry command to see if the scope is running or not. The querry is to return a zero or one. So it is a write and a read.
0 Kudos
Message 4 of 7
(3,278 Views)
It is a read VISA that is timming out. I have tried setting the timeout to 5 minutes but I suspect my change is not affecting code.

Unfortunatly I have to let the instrument trigger on it's own. We are trying to capture intermittant voltage spikes on a power supply line, and have no idea before hand when they will happen. So we set the scope to trigger, and the software polls it to see if it has.

I don't think it is an error buffer problem, because after a timeout error occurs, the current code retries the querry and everything works fine.
0 Kudos
Message 5 of 7
(3,277 Views)
Since your "retry" is a write and read, that indicates the initial write command is not getting through/failing (even if it's the read that is timing out).

Check your NI-Spy capture to see which command is timing out.
-The write should timeout. If it doesn't then the instrument is accepting a command but not processing it and increasing the write timeout would not help If this is the case, then you should contact the instrument manufacturer to report this.
-If the write is timing out then increase it's timeout duration

Ray K
NI Applications Engineer
0 Kudos
Message 6 of 7
(3,276 Views)
Prog Guy wrote in message news:<5065000000050000007FAB0000-1031838699000@exchange.ni.com>...
> The retry retries the entire sub VI, and three different sub VIs have
> the retry ability. The Sub VI that I see failing write a querry
> command to see if the scope is running or not. The querry is to return
> a zero or one. So it is a write and a read.

Prog Guy,

Here's another thought: Would you be using a GPIB-ENET/100 controller
by any chance? If you are, you might need to update the firmware. I
was using rev .6 with this device and it was intermittently dropping
characters and also failing to report NRFD de-assertion to the driver
in Win 2000. This was giving me spurious timeout errors that were
often cured with a re-send of the sam
e command.

gm
0 Kudos
Message 7 of 7
(3,276 Views)