Why does my frequency measurement double
I have a Low-Frequency 1 Counter measurement in a VI that I wrote. There are two of them in the same program, and one of them reads accurately at ~118 Hz. The other one occasionally doubles the frequency. The true signal is ~30Hz, but the display reads 60Hz half the time. What am I doing wrong and how can I fix it?
(3,865 Views)
you may need to look for "aliasing"
(3,862 Views)
To my knowledge "aliasing" occurs to do undersampling a signal, and results in an apparently lower frequency. If the program has no problem distinguishing 118 Hz it is not undersampling 30 Hz. The frequencies stated have been measured by oscilloscope.
(3,859 Views)
Try tracking the delta time between your freq checks. I fmemory serves me that VI accepts a delta time input that if not provided by the caller will be calcualated by the driver based on when it THINKS it was last checked.
Trying to help,
Ben
(3,855 Views)
adding to the above...
a low pass filter (if availabel on borad for your widget) may help if the is a mismatch between the output/input impeadance resulting in reflection or "rining" which could give you multiple counts per unit time.
Ben
(3,846 Views)
That's exactly the issue I had about 15 years ago reading pulses from an instrument. A simple termination resistor at the input eliminated the problem.
(3,834 Views)