[cRio-9039] FPGA data logging
I am trying to log data at the maximum possible sampling rate using a cRIO-9039 with NI-9239 (voltage) and NI-9211 (temperature) modules installed. To achieve high-speed logging, I configured my VI by referring to the method described in the following NI Knowledge Base article:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000x3ruCAA&l=ko-KR
However, I am encountering two issues.
  1. The sampling rate of the NI-9239 is significantly lower than expected.
    As far as I understand, the NI-9239 should support up to 25 kHz when using two channels. In my case, however, the data is being logged at only a few hundred samples per second.
  2. The actual measurement data is not being logged correctly.
    For the NI-9239, the logged values remain at 0, and for the NI-9211, the logged values remain statically at approximately −2, regardless of the applied input signals.
The version of LabVIEW I am using is LabVIEW 2015.
Thank you very much for your time and support.
(1,122 Views)
Accepted by topic author jeongho630
You place both NI-9211 and NI-9239 in the same loop. The NI-9211 is slowing down the NI-9239. Use parallel loops so that both can run at their maximum rate. See the example below to log synchronized data via a single DMA FIFO.
Both loops are from shipping example at Help >> Find Examples... >> Hardware Input and Output >> CompactRIO >> Module Specific IO >> Analog Input
Make sure your signal are connected and grounded correctly. See Troubleshooting Offset, Incorrect, and Noisy Readings of NI Multifunction DAQ
https://tmesystems.net/
(1,088 Views)
Thank you!!
(1,056 Views)