VXI-11 is a protocol for the instrument - not your pc. http://digital.ni.com/public.nsf/allkb/3DD043326CF5BF6786256C10005151DC
Is the IP of your network card set to the same subnet as the IP of the instrument? You need to do that. Are you using a direct cable from your pc to the instrument? If you are, you need to use a cross-over cable.
03-27-2007 09:34 PM
03-27-2007 10:27 PM
03-27-2007 11:06 PM
I put the ethernet cable ( i have tried with a normal and with a cross over). In the spectrum analyser, I set " set the IP adress dynamycally using DHCP ". I have checked on my computer ( which is very recent) that the DHCP is well activated, and it is ok . But the computer tells my that the connexion is limited or inexistant. When I try to use the software provided by the constructor, i can t connect.
Normally when I use DHCP for the spectrum analyser, the IP adress should be displayed after the power sequence. But the only thing which is written is : IP xxx.xxx.xxx.xxx >
Then I have try to put manually the IP adress, but in this case too, it doesn't work. It is very strange
03-28-2007 08:24 AM
03-28-2007 08:48 AM
If you have connected a NIC directly from your PC to the Instrument with a crossover cable you cannot use the DHCP settings. You must be connected to a DHCP server in order to be assigned IP addresses.
On both the PC and Instrument configure the network settings to use static IP Addresses.
Enter an IP Address in the PC manually, (eg. 10.101.0.2) set the Subnet Mask to 255.255.255.0
Enter an IP Address in the instrument the first 3 numbers should match the IP in the computer, the fourth one MUST be different (eg. 10.101.0.5) ensure the Instrument 's Subnet Mask matches the PC.