connecting the NI 9944 to Pasco load cell
I would like to connect teh PASCO PS-2200 Load Cell to the NI 9944. I have performed soldering so that such a connection is possible. Can someone tell me what the EX+ and IN+ are for?  which one is for the positive signal and the negative signal?
(2,796 Views)
Ex+ means the positive side of excitation.  In+ means the positive side of the input reading.
Where did you see these designations, on the NI module or the load cell?
Generally load cells are a full bridge arrangement which means at least 4 wires:  two for excitation, two for sensing.  But you said you have an NI-9944 which is for reading quarter bridge strain gages.
Are you sure you are using the right DAQ device for your load cell?
(2,787 Views)
On page 3 of this Pasco PS 2200 manual ftp://ftp.pasco.com/Support/Documents/English/PS/PS-2200/012-10638A.pdf  you can see the pinouts. Currently I have connected pin 1 of the Pasco load cell to EX + of the NI 9944 and pin 3 of the load cell to IN+ of the NI 9944 following what I understand from the instructions in this link https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019OptSAE&l=en-US  . I then used a jumper wire to connect IN+ to QTR of the NI 9944 by making a sort of hook with one end of the jumper wire so that it can hook to the wire connecting IN+ to pin 3 of the load cell. The Shield pin of the NI 9944 is not connected to anything. Is this setup correct? Pins 2 and 4 of the load cell are connected to a power supply. Maybe I do not have a clear understanding of the terminology used in the instructions.
The DAQ used is the NI cDAQ-9172 and the module is the NI 9237. I believe these devices are all compatible with each other. On page one of the NI 9944 user guide https://www.ni.com/docs/en-US/bundle/ni-9926-9944-9945/page/user-manual.html  it says that a plastic tie strap and some screws are required. What purpose do they serve? Is it really necessary? The goal is to make a quarter bridge. We believe that the signal sensed by the load cell is resistance and so the NI 9237 would make a Wheatstone bridge and the NI 9944 would complete it. Supposedly the readings would be in volts. The signal received by LabVIEW is just a straight line at 0.05 V whether mass is applied or removed from the cell.
Thanks for your help.
(2,773 Views)
That link looks like it is a full bridge load cell.  +4.2V would be equivalent to Ex+.  Gnd is probably Ex-, but generally no corner of a full bridge is going to be tied to ground.  + Strain gage output and - strain gage output would go to In+ and In- on a DAQ card.
However, you have a quarter bridge DAQ device with the NI-9944.  That will not work with a full bridge load cell.  You should just be using the NI-9237.
You should probably contact your local NI sales rep and run through this with them to verify you have the correct hardware for what you are trying to do.
(2,764 Views)