Thanks, lynn. It is a square wave signal.According to the specification of USB6218, the gate of the counter should be fed with the signal to measure, and the source of the counter with a known frequency. Surprisingly, if I connect the source to the ground, the VI reading is the same as when I feed some known frequency to the source. It can also read the correct signal frequency but enormous other mumbers.
09-18-2007 08:24 AM
09-18-2007 11:58 AM
09-18-2007 12:28 PM
09-19-2007 07:28 AM
09-19-2007 07:47 AM
Hi,
your problem sounds familiar to me, if this is what happens:
- you can measure your frequency properly using an analog input
but
- you are getting spurious results using the counter
As already mentioned above, this is a strong indication that there is noise on your signal which the ADC (1 MHz sample rate) doesn't mind but the counter (80MHz) does mind. The counter expects a proper TTL signal with very clean edges - have you checked with a scope ??