09-19-2007 03:20 PM
Have you tried the measurment with max test panels? This will allow you to not blame the code.
Paul
09-20-2007 08:20 AM
09-20-2007 01:21 PM
I have see some weird stuff relasted to CounteInput->General->Properties->Moew->Duplicate counter prevention. This is set by default. Make this false on your task. If you have a low frequency signal and a fast clock whrn you count zero (as the source frequency drops near the clock frequency) the zeros are filtered out and your measurment is incorrect. Aslo check to see if the pulse width is not below the minimum pulse width filter if your device has this option and it is also enabled. Good Luck,
Paul
09-20-2007 02:02 PM
09-28-2007 11:31 AM
10-01-2007 03:02 AM