Hi Jin,
If you are using LabVIEW 8.0, then you need to download DAQmx 8.0 as well. Setting the voltage limits to not include negative readings is not going to filter out that data for you. The easiest way to get only the positive portions of the sine wave is simply to acquire all of the data and then filter out the negative readings in post processing.
However, for NRSE configuration, your connections are correct with the signal connected to AI0 and the ground to AI Sense. Is there a specific reason that you chose this configuration? Here is a link with more information on these configurations: Field Wiring and Noise Considerations for Analog Signals.
I am not sure what switches you are referring to in your second question and I am not sure exactly how you want to trigger your signal. Could you please be more specific?
Regards,
Hal L.