Hi gaetano,
I have a couple of suggestions and a couple of questions.
The first suggestion is to take LabVIEW out of the equation until you can verify that the NI-9215 can read the correct signal. You can use a Test Panel in Measurement & Automation Explorer for the NI-9215 to read you analog input signal. Open Measurement & Automation Explorer (MAX) and find the entry for your device. It can be located by looking under My System >> Devices and Interfaces >> NI-DAQmx Devices >> NI 9215. Right-Click on the entry for the NI 9215 and select Test Panels. You can select a sampling rate and other parameters as needed to test your analog input signals.
Anytime noise large noise levels are observed on a signal it is always a good idea to check for possible grounding issues.
How do you have your signals wired? (differential, single-ended ground referenced, etc)
Is the voltage that you are measuring floating (not connected to ground) or grounded?
Depending on your signal some simple steps may need to be taken. For example if you are measuring a floating signal in differential mode it is recommended to use bias resistors to reduce noise by providing a path to ground for any bias currents.
The following document is an excellent resource for setting up your measurement:
Field Wiring and Noise Considerations for Analog Signals
Take a look at your connections and setup and feel free to post back with the results of your investigation.