error in Volt to Temp conversion Type E
Hi
I think that there is an error in the Volts to Temperature vi in Labview, for type E thermocouple,.... at low temperature.
According to the tables the voltage for a E type thermocouple at 13K should be -0.009797.
the Temperature to Volts.vi in labview gives the correct values.
Now putting -0.009797 V in the Volts to Temperature vi will give 25.8 K
Can anyone confirm this error ?
thanks
N
Edit note : I see in the vi that the conversion V to T is valid for T>70 K (albeit T to V is valid from 4K... as it is the type E thermocouple !)
(4,185 Views)
Hi N,
why don't you read the comments on the blockdiagram inside this VI?
For TC type E it clearly states:
temperature range: -200 C to 0 C: error range:  0.03 C to -0.01 C
So why do you expect correct values at 13K when the formula used is only valid starting from >73K?
Btw. type E thermocouples are recommended to be used at ranges from -200 to +900°C, other sources mention -50 to +740°C! Do you really want to use type E at 13K?
(4,169 Views)
Hi
I read the comments (too late), see my PS note.
For type E thermocouples other sources mention 4K to 400 K, it is actually commonly used in cryogenics at 4K.
N
(4,119 Views)
The Temperature to Volts and Volts to Temperature VIs must be used with a good deal of caution. The temperature and voltage ranges are not the same in both directions for some types of thermocouples.  This is not a problem with the NI implementation but a limitation of the NIST documents from which the VIs were derived.  Read the help files carefully.  Outside the specified ranges the errors can get very large, very quickly.
Lynn
(4,081 Views)
Hi
Yes, you are right. There is no problem with the vi (I just read the "small print" after posting) but the ITS90 polyomials which do not cover temperature below -200 °Cin Volt to Temperarure.vi
I should change the title of this post.
Anyway, I took the data and fitted two more regions and added to my vi.
thanks for ehlp
regards
N
(4,049 Views)