03-20-2016 12:49 PM - edited 03-20-2016 12:52 PM
Hi
I think that there is an error in the Volts to Temperature vi in Labview, for type E thermocouple,.... at low temperature.
According to the tables the voltage for a E type thermocouple at 13K should be -0.009797.
the Temperature to Volts.vi in labview gives the correct values.
Now putting -0.009797 V in the Volts to Temperature vi will give 25.8 K
Can anyone confirm this error ?
thanks
N
Edit note : I see in the vi that the conversion V to T is valid for T>70 K (albeit T to V is valid from 4K... as it is the type E thermocouple !)
03-20-2016 01:00 PM
Hi N,
why don't you read the comments on the blockdiagram inside this VI?
For TC type E it clearly states:
temperature range: -200 C to 0 C: error range: 0.03 C to -0.01 C
So why do you expect correct values at 13K when the formula used is only valid starting from >73K?
Btw. type E thermocouples are recommended to be used at ranges from -200 to +900°C, other sources mention -50 to +740°C! Do you really want to use type E at 13K?
03-21-2016 02:10 AM
Hi
I read the comments (too late), see my PS note.
For type E thermocouples other sources mention 4K to 400 K, it is actually commonly used in cryogenics at 4K.
N
03-21-2016 09:03 AM
The Temperature to Volts and Volts to Temperature VIs must be used with a good deal of caution. The temperature and voltage ranges are not the same in both directions for some types of thermocouples. This is not a problem with the NI implementation but a limitation of the NIST documents from which the VIs were derived. Read the help files carefully. Outside the specified ranges the errors can get very large, very quickly.
Lynn
03-22-2016 12:49 PM
Hi
Yes, you are right. There is no problem with the vi (I just read the "small print" after posting) but the ITS90 polyomials which do not cover temperature below -200 °Cin Volt to Temperarure.vi
I should change the title of this post.
Anyway, I took the data and fitted two more regions and added to my vi.
thanks for ehlp
regards
N