Solved!
Go to solution

Hi. I try to take a data of torque on shaft with strain measurment (half bridge system with 2 gages on 45 degrees). Everytime when turn on offset null data starts from 0 and then jump to 80 Nm. The problem is that jump. I dont have any load to have that jump and i dont have prestressed state. I use filtering ( low pass 1hz) to get over the noises in static system. Can anyone help me with that jump problem. Thanks.

Download All
0 Kudos
Message 1 of 4
(3,319 Views)
Solution
Accepted by topic author stefo83

You use a veery low low pass filter .... so start aquisition, wait for the filter to settle, use DC estinate vi on a long datablock and substract that value from the following data (after filtering, or wait again for the filter to settle......)

 

you will run out of memory

 

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 4
(3,296 Views)

Hi!

 

As Henrik had said, the settling time of your filter is causing the problem. Consider improving the filter and discarding the samples that produce before settling. 

 

Cheers:

Andrew Valko
National Instruments Hungary
0 Kudos
Message 3 of 4
(3,226 Views)

Thanks for reply. I set AC & DC Estimator VI to show me starting and established values. Then with constan I move all values to zero (any time I wаnt). After that I can start loading to the shaft normaly. What you think for that solution?

0 Kudos
Message 4 of 4