is daq bit resolution applied to the range set in daq assistant or is it to physical voltage limits of daq.
Hello, So this is my first post here,
I was wondering about this :
When you set up a new daq with daq assist, is the resolution applied to the voltage / number of inputs you set, or is it a fixed value per volt per channel.
For example :
i have a usb-6015 daq, it has 16 analog inputs, 16-bit resolution 250 ks/s -10 to 10 volts.
so is it like (2^16)/(16 *(10 - -10))= my resolution per channel per volt?
or does this have nothing to do with the ammount of channels,
if i want to measure over 0 to 5 volts, does my 16 bit apply from 0 to 5 or is it still over the daq's scale (which is -10 to 10 volts)
Marco
(2,876 Views)
Accepted by topic author m.caron
You will get full 16-bit resolution on every channel regardless of how many channels are configured.  Most DAQs are multi-plexed so your sampling SPEED specification in usually an aggregate meaning that you can only scan channels at a sample rate = MAX RATE / # of CHANNELS.  Also, most NI DAQs have a Programmable Gain Amplifier on the front end so if you specify a smaller voltage range in your task configuration the amplifier will automatically increase the gain to use the most ADC bits possible.  Check you device spec sheet and manual for details...
LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
(2,873 Views)
This is exactly what i wanted to know, thanks very much!
(2,867 Views)
