issue with PPS signal to trigger AI start
Thank you for your suggestions
@mcduff wrote:
Try to explicitly set the reference clock rate after setting the source. For some X-Cards I believe it is 100MHz, for your DSA card it should be 10MHz. See if that helps.
The PXIe-4468 will only accept 100 MHz, so I set it explicitly but it didn't change anything.
@mcduff wrote:
Lastly, how are you measuring the phase drift?
The analog output generated by the PXIe-4468 goes on a cRIO based PMU (Phasor Measurement Unit).
The PMU is gps synchronized (same GPS antenna connected to the PXI-6683) and it sends via UDP (using C37.118 standard communication) the phase and amplitude measured.
As I compensate the AO filter delay and make sure that the 50Hz pure tone starts on a rollover second, I expect the phase to be pretty close to 0° - and it is the case - but I also expect the phase to be somewhat constant through time - and it's not not.
@mcduff wrote:
One more thing to try, query the output sample rate after setting it; you may not be at exactly 20kHz like you assume.
Good point, I checked and it returns 20kHz.
We have two ears and one mouth so that we can listen twice as much as we speak.
Epictetus
Antoine Chalons
Message 11 of 13
(325 Views)
Let's take a step or two back:
  • Initially the problem was with t0 when you use Analog Input. I assume that has been corrected now. Is this true?
New issue with Analog Out, sample clock seems to drift from your measurements, is that correct?
  • According to the specifications GPS has ±40 ns, <8 ns standard deviation synchronization performance. Assuming worst case performance of 40ns, my quick calculation(please check) gives a max phase uncertainty of ~0.29 degree for 20kHz wave. Not sure what you have on your plot, but are you within specs?
  • I still don't understand the phase delay on the Analog Out, a simulated instrument gives an error when trying to read that property node.
Do you have another GPS, like a Trimble? If so, output a PPS from both the Trimble and 6683H and digitize both on your card. See if both stay in sync for a long period. Doing that, I discovered an unexpectedly long trigger delay of about 400ns when using the PXI trigger line and a NI-5172 Scope Card.
The 6683H should rephase its clock every PPS, but there can be drift in that 1 second time frame where you are doing the measurement. Not sure if you are seeing that.
Message 12 of 13
(313 Views)
Yes.
Yes.
I might be able to borrow a PXI-6289 (multi-function DAQ) to help with that.
Epictetus
Antoine Chalons
Message 13 of 13
(278 Views)