measuring resistance using Labview and sc-2345
I have a temp sender that I need to measure variable resistance as it changes do to change in Temp. I planned on placing it in a water bath and heating it gradually. Recording temperature and resistance as the temp increased. My question is what hardware to use in the SC-2345. The resitance range is 16.1 to 153.9 Ohms.
(3,599 Views)
Is the temperature sensor an RTD? You may want to use the SCC-RTD01 module in the 2345. Also do you have a DAQ card with the 2345?
Hope this helps!
Abhinav T.
Applications Engineering
National Instruments India
LabVIEW Introduction Course - Six Hours
Getting Started with NI-DAQmx
Measurement Fundamentals
(3,586 Views)
Daq 6062e . Temperature sensor a side . How can I measure resistance using the sc-2345?
(3,578 Views)
If you use the RTD01 module, it will allow you to directly convert the resistance to a temperature. The RTD01 provides you with an excitation that will be passed into the resistor and then measured on your DAQ card.
Abhinav T.
(3,565 Views)
I am not trying to read the temperature I am trying to read the resistance. The application i have converts the resistance to temperature on a gage. Differnet temp sensor give out different resistance at different temp. I am trying to match a sensor with the preset gage. So I need to measure resistance.
(3,560 Views)
You should be able to use a NI SCC-CO20 to source a precision current to the DUT (device under test).  You could then use one of the Analog Input modules to read the voltage across the DUT and calculate the resistance.
Resistance = Voltage across DUT / current supplied to DUT
This works because the resistance range you need to cover is within the sourcing capabilities of the SCC-CO20.  Make sure to select the analog input module with the right input voltage range.
You could also put a precision known resistance in series with DUT and apply a known voltage across the two.  Use an analog input to monitor voltage across the known resistance and calculate resistance of DUT.
(3,552 Views)