02-27-2007 06:41 AM
02-27-2007 07:49 AM
02-28-2007 07:11 AM
02-28-2007 07:45 AM
02-28-2007 08:02 PM
Hi sumi,
To start off with, I'd like to say that cFP is asynchronous, so you will not be
able to achieve a sampling rate of exactly 100 ms. You may be able to
achieve a sampling rate of around 100 ms, but it will not always be exactly
that much time between samples.
Have you tried using the Cross Power Spectrum Pt By Pt.vi? Take a look at
that, and let me know if you have any other questions.
Best of luck on your application, and have a great day!!
03-01-2007 10:11 AM
sumi,
In order to accurately measure power, you need to be able to calculate the phase angle and power factor. See this post for more info. http://forums.ni.com/ni/board/message?board.id=170&message.id=171419&view=by_date_ascending&page=1
The example from rpursley8 is probably the best candidate for a cFP.
I think that you will find that the AI-100 does not sample fast enough for what you need to do. You need to be able to take many samples during one ac line cycle. I'm assuming 50/60 Hz. The cycle time for 60Hz is 16.66 mS. A cFP-AI-110 would be fast enough to get you 60 samples per 60Hz line cycle.
If you are stuck with the AI-100, I suggest you look at third party devices that measure the power for you.
03-05-2007 02:21 AM
03-05-2007 02:22 AM
03-05-2007 11:00 AM
03-05-2007 11:39 AM