I can not find 400 Ohm resistor in the store. Can I use 300 Ohm 1/2 W resisitor? check the link http://www.radioshack.com/product/index.jsp?productId=12554233

 

Thanks!

0 Kudos
Message 11 of 18
(1,526 Views)
Solution
Accepted by topic author beechmont

Hi beechmont,

 

as long as you keep Ohm's law in mind you can use any resistor you like...

 

I would suggest to use a resistor with

- atleast 1/4W capability (P=I^2*R, so you could use upto 500Ohm resistors to measure 20mA)

- high accuracy (1% or less)

- maybe you should sort out resistors with bad temp coefficients.

 

Btw. I think the Radioshack website is a very bad designed online-shop. Nearly to no filters for the kind of tool you look for. I cannot even filter for certain resistor value ranges. Gosh... Don't you have any reasonable electronics distributors in the US?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 12 of 18
(1,511 Views)

THANKS A LOT!  

 

I also wonder how to calibrate the transducer. I asked the support of transducer before. They just told me to zero it following the maual. What I want to know how to display pressure reading in labview, not current reading.

0 Kudos
Message 13 of 18
(1,501 Views)

Hi beechmont,

 

so now you ask for some basic math? Like the math tought in school - for me it was in 5th or 6th year?

 

Look up for things like "linear scaling", "offset and gain": it's a simple formula as "y = mx+b"...

 

Or to put it in more clear words:

What kind of calculations are needed to convert a value from 4-20mA to 0-x Pa, when 4mA = 0Pa and 20mA = xPa? Come on...

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 14 of 18
(1,498 Views)

I connect the pressure measure system to the computer and try to configure it. When I open the signalexpress and add a step/DAQ/pressure (bridge). I was told the device is not supported. What is wrong with the wiring or my operation?

0 Kudos
Message 15 of 18
(1,465 Views)

We don't know since you didn't post an actual error message or a diagram of your connections. You can press Alt-PrtScn to capture an image of the error pop-up or just copy the message.  At least post the error number.

 

Can you see your 9205 in NI MAX (Measurement and Automation Explorer)???

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
0 Kudos
Message 16 of 18
(1,456 Views)

I can see 9205 in Measurement and Automation Explorer but it is not identified as a pressure measure device. I guess only module, like NI 9237, can be identified. Maybe I have to choose add step\DAQ\acquire signal\voltage. Then I add a step to process the voltage signal with formula.

0 Kudos
Message 17 of 18
(1,431 Views)

Hi beechmont,

 

NI9205 is a generic AnalogInput module, so yes: you have to create a DAQmx task to read the signal and a DAQmx scaling to convert voltage to Pa...

Best regards,
GerdW