I'm using two daq boards, 6036e and 6731 with a SCB 68 terminal block. If I check the voltage between D GND and PFI line of the SCB 68 terminal block for both daq boards, I can see 5 V difference between them for any GND and PFI lines. so, I can not use the trigger using GND and PFI line. Is it broken or normal?
FYI, so I'm using two PFI lines, which are not biased with 5 V, for the trigger. In that case, the only falling edge works no matter what i choose the rasing or falling edge for the trigger.
Solved! Go to Solution.
05-10-2019 01:13 AM
05-10-2019 05:17 AM
Dear GerdW,
Thank you for your reply. So, the hardware is normal. That is, any GND are referenced with +5 Vdc.(Q, why do they give a +5 Vdc pin more?) If I need real electric ground, I have to couple the terminal block with a real device which sends the trigger signal, not use any GND pin. Am I right?
Thank you for advance
SC
05-10-2019 05:39 AM - edited 05-10-2019 05:40 AM
Hi scchae,
That is, any GND are referenced with +5 Vdc.
No, GND is GND. All other pins are referenced to GND…
Do you know the meaning of "pull-up resistors"?
why do they give a +5 Vdc pin more?
Because that +5Vdc pin can be used as a (small) power supply to external devices.
If I need real electric ground, I have to couple the terminal block with a real device which sends the trigger signal, not use any GND pin. Am I right?
No.
You should connect GND with your external device GND.
There are manuals from NI describing several possible connections!
05-10-2019 09:37 PM
Dear GerdW
Thank you for all your replies. Now, I understood how to measure the signal using DAQ board.
Best regards,