speeding the Keithley 6485 readings with LabView

I'm not quite sure how to change the 'MEASUREMENT' command into ´'CONFIGURE' and 'READ' commands but I assume that's what I had in the attached program, right? 

Also, if I lower the 'Max time' there is no improvement on reading rate. Too low value of 'max time' simply results in fast rate but no readings at all. 

0 Kudos
Message 11 of 20
(4,088 Views)

I found another severe problem in the .vi program: it's not reading the actual current signal from the KE6485 picoameter. It connects with the instrument (I can hear the click) and then read some randomly low current signals (a few nA), even when actual current is as high as 3 mA. 

I tried to change Fetch function to Read, and still same problem. What could have been wrong?

thanks for any help! 

0 Kudos
Message 12 of 20
(4,076 Views)

this problem was solved. 

turns out the 'zero check' needs to be disabled in the vi file.  

0 Kudos
Message 13 of 20
(4,068 Views)

Hi Rodmanhurst,

 

Did you manage to find a way of speeding up the readings at all? I'm facing the same problem! I guess all the previous suggestions didn't speed up the reading enough? I've also noticed that by using just the front panel with no remote control the readings speed up significantly. So yes, it would be good to know where the bottleneck is!

0 Kudos
Message 14 of 20
(3,954 Views)

JKLH: 

no I didn't find a way. max frequency I managed was only 2HZ and I sorta had to live with it.

but post it here if you can find a solution, that'll be a great help to many others! 

0 Kudos
Message 15 of 20
(3,934 Views)

Hi all, I'm doing some automated low-current measurements and I may have some useful information for people in this thread. I don't have 6485, but higher model 6487 picoammeter, though.

 

At any rate, 6487 has internal buffer, which can store several hundreds measured values. It can be programed to do repeated measurements at certain time intervals and store the results in this buffer. Needless to say, access to the buffer is much faster than RS232 or GPIB. In the 6487 LabView driver, there are "multipoint" functions, which can set up these repeated measurements and then read the entire buffer at once.

 

And now about measurement speed: according to the 6487 manual, the instrument can take up to 1000 readings/second. However, such acquisition speed is possible only with low accuracy (resolution), with line filters turned off etc. In the default state, the 6487 has almost all these features turned on and set to achiveve high accuracy. But by turning them down, you can beef up acquisition speed. These things can be changed remotely, too. Of course, you will have to find some compromise between speed and accuracy, as needed in your applications.

 

Hope this helps.

0 Kudos
Message 16 of 20
(3,826 Views)

Hello,

 

I´m also new in the use of LabVIEW and I´ve got the same problems as the previous speakers. Besides I had already called Keithley and told them the problem. They already have advised me to use the internal buffer to get faster measurements. But I´m not sure how to use the special "buffer VI´s" (see the attached files below) to get a accurate measurement. In my file the "Fetch Buffer Data.vi" throws an error message "invalid value for parameter or property" and the circuit broke down. Is it possible to post an example, in which the measurement with the internal buffer is shown?

 

Thanks in advance

SeBaK

Download All
0 Kudos
Message 17 of 20
(3,403 Views)

I have been trying to find some documentation for you.  As you may know this driver is not actually built or supported by NI.  The Keithley 6485 Labview driver is actually developed by Keithley.  I found some example programs, and manuals here. Hopefully, you find them helpful, I don't have the hardware to test anything out.

Jensen
National Instruments
Applications Engineer
0 Kudos
Message 18 of 20
(3,387 Views)

Hello,

 

I hope these help.

The first is a 2000 buffer read of 1000 Current Read and 1000 Time stamp.

The second vi is just a 1000 Current Read buffer.

 

Did these help?

Y.

 

0 Kudos
Message 19 of 20
(3,214 Views)

Hello, I have same problem, readinsfg rate is about 1 reading per 0.4-0.5 sec. I thought the problem lies in the speed of  the rsr-232 interface. But when I'm trying change rate from 9600b to greater values, ammeter always freezes. Anyone encountered this problem? Maybe some special combination of parity, stop-bits, flow control? And, in general, will increasing interface speed cause increasing reading rate?

0 Kudos
Message 20 of 20