What about external trigger ?
I use this command but it seems that I need trigger for each voltmeter, which produce phase error
07-21-2006 08:23 AM
07-21-2006 08:40 AM
Thanks for answering,
I use one trigger for both HP 34401 ei one wire is connected to both, but I still must press button twice to get readings
I can send VI, if it would help
Sasa
07-21-2006 08:49 AM
07-21-2006 09:07 AM
External trigger on HP34401 can be pulse 0-5V or simple switch which connect hot end to ground, I use switch , one for both Voltmeters
If I read good manual for HP, after trigger intialization and read command, multimeter should go to the "wait for the trigger" state and when this trigger occurs it should transfer data into buffer, internal or external, so that I can download this data.
I have paralellel thread for both instrument, it seems ther is some problems with RQS.vi ie how to handle one for both
I have to use for low frequency(<5 Hz) AC measurement.
VI is attached, thanks in advance
07-21-2006 09:44 AM
02-27-2009 05:05 AM
Hi.
I have one question to ask.
What will happen or is it possible to send through NI-488.2 command "Send" TARM SGL to both instruments/DMM's at the same time?
Thanks.
02-27-2009 05:25 AM
I'm using two HP 3458A .
Also, how about to set TARM AUTO, TRIG SYN, NRDGS 1, TIMER/SYN?
Thanks,
Kristina.