How to set debounce filter time

Could anyone tell me how to set the debounce filter for the digital input Port 1/ Port 2 of the NI 6259 using LabWindows/CVI 8.1.1?  According to the specification sheet, the choices are 125ns, 6.425us, 2.56ms, and disable, but I couldn't find what function to call to select these options.
TIA
0 Kudos
Message 1 of 2
(3,168 Views)
Hi nantek,

You will need to set the Digital Filter Enable to true.  You must also configure the Digital Filter Minimum Pulse Width attribute/property. This value represents the minimum value that is guaranteed to pass into the TIO. The minimum pulse width guaranteed to be blocked is one-half of the Digital Filter Minimum Pulse Width attribute/property.

This is discussed in the knowledge base article, Enabling the Digital Filters for Counter/Timer Devices in NI-DAQmx.