Proportional shift in frequency of sampled data?

I am experiencing strange behavior with my LabWindows program (using the most recent versions of LW and NIDAQmx). I have two separate threads: 1 for reading 16-bit input from a PCI-6120 board, and a 2nd thread for writing 16-bit output to the same PCI-6120 board. For test purposes I read 1-second worth of samples at a time, and then write those samples (without any processing) to the output channel. Depending on the sample rate that I use in my call to DAQmxCfgSampClkTiming(), I will get a proportional shift in frequency for the sampled data. If I input a sine wave with frequency F0, I will either get a sine wave with the same frequency *or* I will get a sine wave with frequency 0.985xF0. In other words, the frequency will shift down by 1.5 %.

I am using a function generator to input the sine wave, and a Stanford Research Systems SR785 spectrum analyzer. This table shows the relationship between sample rate and the frequency of the sine wave that is outputed when I input a 10 kHz wave. The second table shows the same, but for the case when I input a 1 kHz wave.

Fs (kHz) | F0 (kHz)
----------------------
297 | 10.00
298 | 10.00
299 | 9.85
300 | 9.85
301 | 10.00
302 | 10.00

Fs (kHz) | F0 (kHz)
----------------------
297 | 1.000
298 | 1.000
299 | 0.985
300 | 0.985
301 | 1.000
302 | 1.000

I see this 1.5 % shift whenever I use a sample rate that is near one of the (20 MHz / x) exact sample rates that the PCI-6120 actually uses. For the data in the above tables, the nearest corresponding exact sample rate is 20 MHz / 67 = 298.507 kHz or 20 MHz / 66 = 303.030 kHz. But even though when I request a sample rate of (for example) 299 kHz, I actually get a sample rate of 298.507 kHz, that shouldn't produce a shift in frequency in my data because the same sample rate is used for input and output.

Does anyone have any ideas on what could be producing this frequency shift?
0 Kudos
Message 1 of 4
(3,317 Views)
Hi Eric-
 
My first inclination is that the frequency shifting might be less related to the integer divisor setting, but rather that your sample section size and sampling rates are integer multiples of each other when you see the problem.

Do you see the shifting at different regions of the update/sample clock rate? If so, can you elaborate on that a bit more? Also, are you simply setting the sample clock rates to the same value or are you exporting one clock to the other subsystem for synchronization? If you wouldn't mind paring down your code so I can take a glance at how the data is read and sent to the output I would appreciate it.
Thanks a lot-
Tom W
National Instruments
0 Kudos
Message 2 of 4
(3,293 Views)
After speaking with Daniel (and I think he consulted with you Tom) we discovered the problem and solution. Apparently, even though I request the same sample rate for the input and output DAQ tasks, because they use different circuitry to derive their clocks from the 20 MHz timebase, I end up with slightly different sample rates. This means that instead of my "straight-through" system being: Output(t) = Input(t), I end up with Output(t) = Input(t x alpha), which is equivalent to Output(f) = (1/alpha x f). That frequency scaling is what produced the shift in frequency of my sine wave. If I had used a broadband signal as my input then I would have seen it's bandwidth narrow on the output.

Solution: For the input task I use the default ("") source terminal in my call to DAQmxCfgSampClkTiming(). But for the output task I specify the analog input clock ("ai/SampleClock") so that the same clock is used for input and output. Note that you get an error if you specify "ai/SampleClock" for both the input and output.
0 Kudos
Message 3 of 4
(3,277 Views)
Hi Eric-
 
Great!  That's exactly what I was getting at in the second paragraph of my post.  Regardless, I'm glad to hear you're up and running now.  Please let us know if you run into anything else.
 
Thanks-
Tom W
National Instruments