LabWindows/CVI
USB-6009 coefficients for scaling
Hi, I'm currently using USB-6009. I'm using DAQmxReadBinaryI16 to read data and scale the binary data to volts. I use the function DAQmxGetChanAttribute to get the DAQmx_AI_DevScalingCoeff for each channel, I am using Single ended Analog input of 8 channels. However, I found the offset of my waveforms far below zero, I suspect that
the scaleCoef[0] which is far below zero that caused it.
scaleCoef[0] = -31.157441
scaleCoef[1] = 0.000638
scaleCoef[2] = 0.00
I am using the polynomial functions of ax^2+bz^1+cx. Is this correct?
I feel that the scaleCoef[0] should read a very low value below 1.0. Which function should I use to get the correct scaleCoef. Please advice.
(3,330 Views)
Hi Kling,
 You can use the DAQmxGetAIDevScalingCoeff function to retrieve the device's scaling coefficients.
The polynomial you used is not correct, it should be the other way round ie. ax + bx^1 + cx^2...
You can refer to this KB for more information on scaling:
Is DAQmx Raw Data Calibrated and/or Scaled?
Best Regards,
Boon Chen
NI ASEAN