LabWindows/CVI
VISA ERROR:1073807195 CVI can´t control 82350A GPIB card
I have a problem with an agilent 82350A PCI/GPIB card,
CVI is reporting ERROR 1073807195 "The interface type is valid, but the specified interface number is not configured".
I´ve checked that NI-VISA is primary VISA library (C:\windows\system32\visa32.dll --> right click \ properties \ version TAB).
I´ve also checed in MAX \ Tools \ NI-VISA \ VISA Options \ Passports --> NiVi488.dll -- NI-VISA Passport for GPIB is checekd.
I can see the card runs OK in Device Manager (GPIB INTERFACES --> Agilent Technologies 82350 PCI GPIB) and in Agilent Connection Expert (GPIB0 --> all instruments GPIB0::13::INSTR ....) where i can check that the target is ok because I can communicate with instruments...
I must install Agilent IO Libraries 15.0 (this install the driver for the card agt82350.sys)
Under MAX I can´t see the GPIB card.
Any help?
Thank you so much
Félix
(4,421 Views)
Accepted by topic author norak
The solution to this problem is that the primary VISA must be of Agilent, not of NI.
this fix the problem, CVI doesn´t report any problem (although it is using agilent VISA library...)
thank you anyway.
(4,395 Views)
