I need to use both trigger-in and trigger-out events. It appears there is only one front panel port, i.e. SMB PFI<0>. If I use that port for the trigger-in, what options are there to get a trigger-out signal out of the box (not pxi bus)?
Solved! Go to Solution.
10-27-2011
04:09 PM
- last edited on
08-24-2025
04:44 PM
by
Content Cleaner
The PXI 6541 actually has four PFI lines which can be used for importing and exporting triggers - though as you've determined, only PFI 0 has a direct front panel connection. The others (PFI 1-3) are accessible through the Digital Data & Control port, and thus can be easily accessed when using either the CB-2162 or SMB-2163 connector block (or any similar breakout for the DDC port):
https://www.ni.com/docs/en-US/bundle/cb-2162-user-guide/resource/cb-2162-user-guide.pdf
https://www.ni.com/docs/en-US/bundle/smb-2163/resource/smb-2163.pdf
Hope this helps!
James M. | Applications Engineer | National Instruments