read lvdt in MAX

I'm using " measurement & automation explorer" for the first time. I'm trying to aquire a reading from AC LVDT sensor. In analog input configuration menu, I'm prompted to give exication souce voltage, sensitivity units, range?
My questions are:
exication source voltage, is that in VRMS of AC peak to peak?
sensitivity units: in specs for the sensor sensitivity is given as 4 V/inch/V, but in MAX only option with inches is mV/V/0.001in. The units seem to me different....? Should I input 4 mV/V/0.001in?
and finally, range: is that linear range of the sensor or stroke?
 
I've tried to find answer to these questions in the help file, but I couldn't find anything.
Please let me know,
Thanks,
Ivan
 
0 Kudos
Message 1 of 2
(2,894 Views)
Hi Ivan,

In your post you mentioned that you are using MAX and trying to aquire some LVDT measurements for the first time, but you had a few questions:
  1. Excitation Voltage:  The Excitation Voltage refers to the V_RMS value of the AC excitation voltage.
  2. Sensitivity Units: The only values supported are mVolts/Volt/0.001 Inch and mVolts/Volt/mMeter.  In your 1 mV/V/0.001 inch = 1 V/V/1 inch (0.001/0.001 = 1).
  3. Range: Refers to the maximum linear range for the sensor.
I hope this information helps you.