Dark Pixel Line In Captured Image

Solved!
Go to solution

Hello,

 

I am trying to use a high speed machine vision camera (AVT guppy PRO F503C) to analyze touch screen LCD user interface modules for pixel defects, foreign particles, scratches, etc... However, this process is hampered by a dark line of pixels which tracks down the screen continously during image acquistion (both grab and continous grab). I have had a little success adjusting the camera shutter value to minimize the visibility of this line, however, this has proved to be relatively inconsistent as the shutter value seems to differ from unit to unit. I am wondering if anyone has ever run into this issue and if there is any corrective action to eliminate or minimize the effects of this line (lighting, camera settings, etc..). Here is an image:

 

dark pixel line.png

 

Thanks,

Chase Fearing

Test Engineer

Subzero / Wolf

Chase Fearing
Test Engineer, Certified LabVIEW Architect
0 Kudos
Message 1 of 11
(6,154 Views)

-hardware settings-

firewire communication

pixel format: mono 8

speed: 400mbps

Packet: 2048

 

-software-

imaqdx

vision development

Chase Fearing
Test Engineer, Certified LabVIEW Architect
0 Kudos
Message 2 of 11
(6,146 Views)

You need a camera with Global Shutter.

Machine Vision, Robotics, Embedded Systems, Surveillance

www.movimed.com - Custom Imaging Solutions
0 Kudos
Message 3 of 11
(6,143 Views)

To add a little explanation to MoviJohn's response, that camera's sensor exposes a line at a time rather than all at once (a rolling shutter). This causes problems if either your lighting has flicker (like from incandescent or flourescent lights) or the part is moving while exposing.

 

Assuming your part is not moving, you would just need to make sure that your lighting does not flicker. If you removed your existing lighting sources and added some constant lighting (like from LEDs) then you could use a rolling shutter. Moving to a global shutter camera could help, but this could just make you get inconsistent lighting frame-to-frame if your light source is still flickering during the exposure (unless you always expose significantly longer than the flicker rate).

 

Eric

Message 4 of 11
(6,137 Views)

Eric,

 

Thanks, currently the test fixture is exposed to flourescent ambient light, but I get the same results when the lights are turned off. I've experimented with different spot lights, bar lights, etc.. but have not come across a solution. Do you think it's possible the flicker is emmitted from the LCD screen itself?

 

Thanks,

Chase

Chase Fearing
Test Engineer, Certified LabVIEW Architect
0 Kudos
Message 5 of 11
(6,134 Views)

Ah, from the image I thought it was like a label but now I re-reaad your post and see that it is a screen. Generally those have backlights that flicker and the LCD itself refreshes at specific intervals. I don't have any experience imaging such a part but my guess is that if you crank up your exposure time (and reduce the aperature of your lens so it doesn't over-expose) such that it covers many refreshes then it seems like it would hide it.

 

Eric

0 Kudos
Message 6 of 11
(6,132 Views)

Here is a workaround.  No charge.  This time 🙂

 

I'm assuming the line always shows up in a different position vertically in the image.  If this is not the case, you may need to change the framerate so that the line moves around.  The VI I have attachec take in images from your camera until the "# of Images to Consider" value is met.  While it is collecting the images, it builds a sigle array of the max pixel values that came before.  The resulting image will not have the line.  You will need to play with the # value to figure out what works for your application.

Machine Vision, Robotics, Embedded Systems, Surveillance

www.movimed.com - Custom Imaging Solutions
Message 7 of 11
(6,122 Views)

MoviJOHN,

 

That is very generous of you. Is there any chance you could post a LabVIEW 2010 version of the VI?

 

Thanks,

Chase

Chase Fearing
Test Engineer, Certified LabVIEW Architect
0 Kudos
Message 8 of 11
(6,106 Views)

I think there is a forum where you can post conversion requests.  Upload it there.

Machine Vision, Robotics, Embedded Systems, Surveillance

www.movimed.com - Custom Imaging Solutions
0 Kudos
Message 9 of 11
(6,092 Views)
Solution
Accepted by topic author CFEARING01

Oh!  I didn;t know I could do that!  Here it is in 2010 format.

Machine Vision, Robotics, Embedded Systems, Surveillance

www.movimed.com - Custom Imaging Solutions
0 Kudos
Message 10 of 11
(6,086 Views)