09-18-2006 11:25 PM
Did someone have experience to use downloaded driver for SR 245 to take data? I used GPIB-USB-HS cable (from NI) to collect data, and corrected GPIB address in the driver, which is found by M&A Explorer. When I ran this program driver, I saw ERROR light flashing one time and Activity light flashing for another time, but I got nothing for data. I used BNC cable to connect input signal in INPORT 1 in the front panel of SR 245, so there should have signal input. Do I need to change something in the program?
Thanks
Linghui
09-19-2006 02:26 PM
09-20-2006 01:34 PM
Thank you.
I used Max to communicate with the instrument, but it doesn't work. When I sent command ?1, and query, the string recevied is iberr=EABO. When I enter other commands, the Activity LED also blink quickly.
When I ran downloaded driver for SR 245, Acitivity LED and Error LED blinked. I didn't receive any error message, and I didn't receive any output data.
09-21-2006
12:21 PM
- last edited on
11-11-2025
05:40 PM
by
Content Cleaner
Hi Linghui,
The EABO error usually indicates a timeout. This could be caused by a few different things such as sending an incorrect command or not appending the correct termination character. I would check the manual for your device and see if it requires a certain type of termination character, as well as what commands it accepts. You can also find troubleshooting techniques for this error here and here.
Hope this helps!
Adam W
Applications Engineering