07-07-2005 01:57 PM
07-08-2005 03:50 PM - edited 07-08-2005 03:50 PM
Hi Jason:
First to address the instrument that does not show up at all when you scan for instruments. I'm assuming since you used this instrument with your ISA board that it could be pretty old. Sometimes the scan for instruments will not work with these older instruments, however you can still communicate with them using GPIB commands. One way to check this is through the interactive control. This can be done by right clicking on your GPIB board in MAX then selecting Interactive Control. Attached to this e-mail is a screen shot of what interactive control should look like. The only difference you will want is possibly a different command in the event your instrument does not respond to *IDN?.
As for the instrument that is not coming up with the proper address. Unfortunately I do not know why this is happening, or how to fix it. With this since scan for instruments does find the instrument you should try to open a VISA session to it. Right click on the instrument and select Open VISA Session. VISAIC will open and you should select the Basic I/O tab. Execute a write then execute a read. If you write *IDN? and get the instrument description when you read, then we have verified that we can communicate with the instrument. If this does not work I would recommend that you try interactive control as you did for the previous instrument.
Regards,
Emilie S.
Message Edited by Emilie S. on 07-08-2005 03:51 PM