RTD Temperature Measurement with NI 9217

I have used a NI 9219 in combination with some RTDs to measure temperature in the past.  I configured the channels using the code below.  I am trying to use a 9217 now, but it doesn't seem as though the configuration is the same.  When using the same code as before, it returns NaN.  When using MAX, it returns the resistance.  The 9219 would return a temperature.  How do I configure the channel for an RTD using the 9217?  What do I need to do to get temperature from the value returned by the 9217.

 

Thanks

 

 

Case Is = "3750" Tasks(TaskCount).AIChannels.CreateRtdChannel(.Device & "/" & .Channel, "ai" & .Device & .Channel, _ Convert.ToDouble(.Min), Convert.ToDouble(.Max), _ AITemperatureUnits.DegreesC, AIRtdType.Pt3750, AIResistanceConfiguration.FourWire, _ AIExcitationSource.Internal, Convert.ToDouble(0.001), _ Convert.ToDouble(.Report))

 

 

Programming Data Acquisition and Control in Measurement Studio and Labwindows/CVI
0 Kudos
Message 1 of 5
(6,432 Views)

Hi Michael,

 

It appears you are looking to do a 4-Wire RTD measurement using the 9217.  Also, it seems you were able to successfully do this using a 9219 before using the same syntax, is that so?  When you refer to reading in the value in MAX as a resistance, I am a little confused.  I know that it is possible to configure a DAQmx RTD task in MAX for this module that will return a value in DegC.  Have you tried this in MAX?

 

 

Here are a few documents to review about the physical connections:

 

How Do I connect 2,3, and 4 Wire RTDs to My Data Acquisition Card?:

http://digital.ni.com/public.nsf/allkb/5BFB36CFCC32C662862570D5006DC0C0

 

The NI 9127 and NI 9219 2-Wire RTD Support:

http://digital.ni.com/public.nsf/allkb/09E99F60C481D613862572C300637341

 

 

It also appears that you are programming in .NET.  Are you using VB.NET or C#.NET?  We do provide a C reference help for DAQmx, and the syntax is very similar to that of .NET.  It can be found by navigating to Start»Program Files»National Instruments»NI-DAQ»Text-based Support.  There are also Examples for .NET located here.  The topic I searched for in the C Reference Help was "DAQmxCreateAIRTDChan."

 

Also, the examples folder as mentioned above contains a shipping example titled ContAcqRTDSamples_IntClk.  Within this folder you can find an example for C# (CS) and for Visual Basic.  Please try out this example to see if it works for your application, and then you could use the code for the task setup in your code.

 

Here is the code from the c# example:

 

myTask.AIChannels.CreateRtdChannel(physicalChannelComboBox.Text,"",

Convert.ToDouble(minimumValueNumeric.Value), Convert.ToDouble(maximumValueNumeric.Value),

AITemperatureUnits.DegreesC, rtdType, resistanceConfiguration, excitationSource,

Convert.ToDouble(currentExcitationNumeric.Value), Convert.ToDouble(r0Numeric.Value));

 

Here is the code for Visual Basic:

 

 myTask.AIChannels.CreateRtdChannel(physicalChannelComboBox.Text, "", _

Convert.ToDouble(minimumValueNumeric.Value), Convert.ToDouble(maximumValueNumeric.Value), _

AITemperatureUnits.DegreesC, rtdType, resistanceConfiguration, _

excitationSource, Convert.ToDouble(currentValueNumeric.Value), _

Convert.ToDouble(r0Numeric.Value))

 

 

Please let me know if I have misunderstood or overlooked anything in your post.

 

Best,

 

 

Adam
Academic Product Manager
National Intruments
0 Kudos
Message 2 of 5
(6,414 Views)

It is possible that the engineer hooked up the 4-wire RTD incorrectly.  It was returning ~109-111 using MAX.  If I rubbed it it would increase, but not by much.  I don't see a place in max to specify the RTD type or R0, however, so I'm nore sure what it returns and how calibrated that value is.

 

In what case would DAQmx return NaN when trying to read an RTD?  I got my code from the Measurement Studio samples, and it looks just like what you have provided.

Programming Data Acquisition and Control in Measurement Studio and Labwindows/CVI
0 Kudos
Message 3 of 5
(6,408 Views)

Hi Michael,

 

I just finished simulating a cDAQ 9172 chassis with a 9217 in slot 1.  I then proceeded to use the Visual Basic Example that was mentioned before in my post.  I had to change the current excitation source to be internal and the value to be 1mAmp.  This was because it threw Error -200077: this value is not supported.  After changing those values, I was able to receive working values back from the VB Main Form splash screen.  

 

With that said, the NaN could be coming from the external connections not working correctly.  If you could, please try to simulate a cDAQ chassis (9172, 9174, 9178) in MAX and then configure the chassis to have a 9217 in slot 1 of the chassis.This can be done in MAX by right-clicking on Devices and Interfaces and selecting Create New...»DAQmx Simulated Device»9172, 9174 or 9178.

 

You can then create a DAQmx Task in MAX by right clicking on the cDAQ 9217 Device that is now listed underneath the cDAQ chassis and selecting Create Task....  Then proceed to create an Analog Input»Temperature»RTD task, and configure it accordingly.  Once you get the task created, you should see a window similar to the image below where you can change the RTD type, excitation type, etc.

 

rtd.png

 

Let us know how it goes after simulating a device in MAX.

 

Best,

Adam
Academic Product Manager
National Intruments
0 Kudos
Message 4 of 5
(6,403 Views)

Bad programming on my part.  I grabbed an available variable to use for R0 and that variable's type happened to be boolean.  A 1 or -1 was not a proper value for R0, hence the NaN.  Problem solved, and it works with the same code that I used on the NI 9219. 

 

Thanks for your help.  Sorry about the waste of time!