As the PCI-7334 is a motion control device, it's measurement capabilities are very limited. Of course you can read encoder values and analog inputs in NI-Motion, but the timing is done in software and it is non-deterministic. That means, there will be always a delay between your readings and you can't tell the exact amount of this delay, which might be in the magnitude of 5 to 10 ms.
For measurement tasks better use an M-Series board like the PCI-6220. These boards offer hardware timed acquisitions and you can correlate encoder readings and analog readings very accurately (syncing accurracy is in the lower nanoseconds). If you use the internal RTSI connector of the PCI-7334 and the M-Series board, you even could route the encoder signals from the motion controller to the M-Series board without the need of external signal connections.
I hope this helps,
Jochen Klier
National Instruments