04-16-2008 02:18 PM
04-16-2008 02:57 PM
Gordon:
Is the negative reading happening with the DUT by itself, unpowered? Are there any capacitors in the circuit that could be storing a voltage to cause an erroneous reading?
Are the probes connected properly at both ends in terms of polarity?
What reading do you get when you short all 4 probes together by themeselves?
04-17-2008 01:27 AM
04-17-2008 10:37 AM
04-17-2008 02:14 PM
04-17-2008 03:29 PM - edited 04-17-2008 03:31 PM
04-18-2008 01:10 AM
Hi,
Maybe it is reassuring but that doesn't say that it is good !!![]()
Still depends on what you want to measure. You talk about 'connector and ground bolt' it is possible that you are using the wrong method. On what standard should the test be based ?
04-25-2008 07:39 AM
Hey guys, sorry for the late reply
I'm not too familiar with measurement standards, but there are areas where the test procedure asks to measure between the frame and a particular bolt to make sure the resistance is under 7 milliohms. The test document also asks that the equipment to have accuracy of 2% of full scale. If I'm using the PXI-4072 calibrated to the NIST standards.
In another topic, does anybody have any idea what negative readings mean, or are they just anomalies?
04-27-2008 04:27 PM - edited 04-27-2008 04:28 PM
04-28-2008 01:47 AM
Hi,
I agree fully with Nitad.
If you are testing a ground connection it must be possible to test with a high current. Testing safety earth connections must be performed with at least 25A (Europe) With these high currents it is much easier to measure the voltage (resistance).
Currently I am looking for a resistance meter for crimp contacts and I can asure you it is not easy. They (the manufacturers) all say that their equipment can do the job better than the other one. Crimp contact resistance must be measured with a voltage not higher than 20mV and a current not higher than 100mA. We are talking about a resistance of 0.5mOhm !!!
So it is possible to measure a small resistance with a small current but it is not easy. One little mistake can give you the reading you get. Go through the whole test setup again step by step. You already said that shorting the test clamps give a reading near zero. So it must be something from your DUT that is causing these false readings.