01-02-2007 03:38 PM
01-03-2007 08:53 AM - edited 01-03-2007 08:53 AM
Message Edited by ampm on 01-03-2007 09:02 AM
01-03-2007 02:41 PM
01-03-2007 04:26 PM - edited 01-03-2007 04:26 PM
Dear Malay Duggar,
Thanks for your reply!
I had the reference position for the acquisition set to 0 (in "niScope
Configure Horizontal Timing ") before. I also added the "Reference
Position" property node as you said. This does not help. The AWG is still
delayed by the same 40 samples... I can not find how to set the number of
pretrigger samples for the digitizer to zero...
I think the problem is not related to the pretrigger samples of the Digitizer because the delay of the AWG does not depend on the aquisition rate of the Digitizer. It depends on the sample rate of the AWG and is always ~ 40 AWG samples long.
I hope you can get a general idea about how I program AWG and Digitizer by taking a
look at the couple of pictures attached.
Generally: I have a Stacked Sequence structure in which one of the frames
contains tasks shown in the "frame 1" figure and the next frame
("frame 2" figure) contains a "for" cycle in which I start
and stop tasks from the "frame 1" repeatedly. LabView code shown on
the pictures is part of the larger application. I can extract the "trigger
+ digitizer + waveform generator" part of the application into a separate VI
if you need. Hopefully the information from the screenshots will be useful as
well and will help to resolve the problem in hand...
Thanks!
Message Edited by ampm on 01-03-2007 04:30 PM
01-04-2007 12:28 PM - edited 01-04-2007 12:28 PM
Message Edited by ampm on 01-04-2007 12:34 PM
Message Edited by ampm on 01-04-2007 12:35 PM
01-04-2007 05:03 PM
01-05-2007 04:10 AM
01-05-2007 10:33 AM
Hi 50ohm Terminator,
Thanks for the information! It definitely explains the situation I have.
RTFM is really helpful sometimes 🙂
What are the technical reasons and justifications for such (I would say unexpected)
behavior of the 5412 (and I guess other) signal generators?
Thanks for your help!
01-05-2007 10:46 AM
Hi Partha,
Unfortunately I was not able to save the VI for
I am using
"C:\Program Files\National Instruments\LabVIEW
8.2\instr.lib\niFgen\niFgen.llb\niFgen Abort Generation.vi
Cannot save VI from a toolkit to a
previous version of LabVIEW."
during the conversion.
Hopefully somebody else can help you to make the VI run
with
Best,
ampm
01-08-2007 11:23 AM