Hello Glen,
The first thing that comes to mind is that this may be noise from the power outlet (60 Hz). Now, the question is how are you getting it?
The first thing I would like you to try is perform your acquisition with everything internal. i.e. no external scan clock and check if you still see the 60 Hz signal. If not, then you are sending it in somehow with your wiring. I doubt though that this is the case. I am leaning more toward some kind of noise introduced by the device you are measuring from. You said that when you selected Floating Source from the BNC-2110, you had a smaller noise amplitude. I would recomment setting it back to GS and using bias resistors are described in the following tutorial..
Also, you said you connected the scope to an analog input line. During your acquisition, you did not see the 60 Hz noise you said. That means it is not really affecting your acquisition. If this is the case, then you should not worry about it.