06-02-2008 12:53 PM
Solved! Go to Solution.
06-03-2008 09:00 AM
CHS,
It sounds like the output is latching up. A lot of times this is caused by electrical noise or voltage spikes. Can you describe the heater and the connections between it and the DAQ device?
06-03-2008 10:10 AM
Thanks for the responce, the heater control is a Watlow DINAMITE-C SCR with the 0-10VDC option. The cabling is about an 18 or 20 guage 2 conductor shielded wire of about 20 feet, until it gets into the box with the SCR mounted, it is not really near any high voltage high current lines.
CHS
06-03-2008 10:33 AM
CHS,
SCR in Watlow could generate some nasty hash that may be getting into the 6211. Two ways to get in, power supply and connection between DAQ output and Watlow.
Since the problem usually happens just to the channel driving the Watlow, I would start there. You said it was shielded wire. Can you verify how the shield is currently connected?
06-03-2008 11:50 AM
12-12-2008 11:59 AM
The wire was installed but the shielding was not connected, that was connected and we have been running for quite some time with no problems so it seems to be solved. Thanks for the input
12-12-2008 01:46 PM
Centerbolt:
Good remote troubleshooting advice there. Kudos to you!
"Captian, shields are back online..."
-AK2DM
12-12-2008 02:03 PM
CHS,
Glad you got the problem figured out. Hope you enjoyed your visit to the UK.
AK2DM,
Thanks. It seems I remember the painfull lessons from the past the best.