9205 - Differential channels for grounded signals and floating signals
for my application I need to use the analog input module 9205 to measure both grounded signals and floating signals.
The 9205 is used with a cRIO 9057 which is supplied by a 24Vdc power supply.
Some signals to be measured are floating (two wires) and the 9205 should be configured in differential mode. Other signals are single-ended, referred to the 24Vdc power supply "-" terminal (GND). Is it possible to use GND as a second wire of the differential pair and measure the single-ended signals using differential channels?
Thank you
(1,330 Views)
You can still configure a single-ended signal to be measured differentially as long as you abide by the instrument's common mode voltage and differential voltage limits.
Santhosh
Soliton Technologies
(1,306 Views)
Hi edomar,
Measuring single-ended signals with differential channels is the preferred method. Take a look at Table 1 on our white paper.
Depending on your setup, measuring a floating signal source in differential mode can present issues with input bias currents on the DAQ input causing the measurement to float out of range.
Alex
Hardware Engineer
(1,298 Views)
Thank you for your reply!
(1,284 Views)