Greetings YZ,
First you will want to make sure that the device passes the Self-Test in MAX. Open MAX, highlight your device, and first click the Reset Device button, then the Self-Test buton (ensure both are successful). After that, test to see if the offset is still present. If so, it sounds like your board needs to be calibrated...there are several ways to perform the calibration. The easiest would be to perform a self-calibration on the device. This can be done in MAX or LabVIEW. In MAX, right-click on the device and choose Self-Calibrate. From there, follow the instructions in the wizard.
If these do not eliminate the offset, we might need to take the calibration a step further by using Calibration Executive or sending the device to NI for internal calibration. Please let me know if these options do not eliminate the offset...
Regards,
Nicholas B, National Instruments
11-20-2006 10:40 PM
11-21-2006
10:02 AM
- last edited on
07-11-2025
09:28 AM
by
Content Cleaner
Greetings YZ,
Glad to hear that you've narrowed down the problem. If it persists, please do not hesitate to post again. The most current version of MAX is 4.1 and can be obtained by installing any of our current NI drivers, available here.
Regards,
Nicholas B, National Instruments